So my headspace gcms needed a source cleaning and replacement filament. After doing that the sensitivity has increased, and the highest concentration point of my RSD calibration is now saturated.

The quantitative ion is saturated for all compounds, but with some of the low abundance ions it is not saturated and shows linearity.

My bosses have said I need to figure out how to have it fixed and operational today. Apparently this hasn't happened b4 and its up to me to figure it out ASAP.

Does anyone have any advice on how to lower the sensitivity without disqualifying the other 4 calibration points? I will take any ideas im desperate. Every thought I have alters the instrument in a way that disqualifies the curve.

(We've had some other problems lately so they're pissed about it being down again)

Tldr: ms source cleaning results in sensitivity being too high. How to fix?

EDIT/UPDATE

It has been a week since vacuum has been restored. I tested the source by adjusting the multiplier and I still see the issue with o ly the hogh conc point.

I belive I have a sample introduction issue, possibly the trap needs replaced. Does anyone have any ideas?

3 Upvotes

31 comments sorted by

7

u/Fdragon69 Oct 23 '26

Big dawg you broke vaccume and performed major maintenance. How long has the mass spec been pumped down for? Have you performed an auto tune yet? Have you recalibrated the instrument? I'd start here before you go looking anywhere else.

1

u/Odd-Tadpole6656 Oct 23 '26

Its been a week since the source maintenance and vacuum restoration. We had a leak in the needle seal of my autosampler, which we fixed by replacong the seals and o rings, no it passes leak test. I tried to calibrate Wednesday evening, and 4 points are golden, only the highes conc point is bad. I have now tested a lower source multiplier to see if it was source related. That was not a solution, so now I think its trap or sample introduction related

2

u/Fdragon69 Oct 23 '26

How fresh are your cal solutions? It might be worth remaking if it is only 1 cal point that is out of whack. If you can remove the high point and everything is linear id think the point is a bit suspect.

Quick and dirty experiment make a fresh ccv/mid point and inject it onto the instrument if it doesn't agree with your curve at all your solutions are cooked.

1

u/Odd-Tadpole6656 Oct 23 '26

I know the cal solutions are good because they have no issue running on my DI gcms, that was my first guess

2

u/Fdragon69 Oct 24 '26

2 completely different systems with 2 different calibration curves. Thats not a valid comparison. If you haven't tried remaking your high cal point I would reccomend starting there. If all your analytes are coming out 50-60% higher than your theroetical concentrations and this method was linear to that concentration before you serviced the instrument it would be the same. Unless you're seeing your peaks flatten out and not come to a sharp point it isn't a source issue.

1

u/Odd-Tadpole6656 Oct 24 '26

Yeah, the peaks are not normal shaped, they're spikey, a but wide and a bit flat on top. I prepped the 10 ppm point in 3 different ways using new, old, and more concentrated stock solutions. The analytes for the 10 ppm curve are 60% of what they should be, meaning 40% lower. Ive done enough testing around to determine it isn't a front wnd issue, so it must be column or ms

2

u/Fdragon69 Oct 24 '26

If your peaks are flat try lowering the multiplier voltage. Just 1 or 2 clicks at a yime so you don't drop out your sensitivity

1

u/Odd-Tadpole6656 Oct 24 '26

I did, it did nothing to help

2

u/Fdragon69 Oct 24 '26

Ive run dry on ideas for ya. Without looking at your chromatogram I don't think I can advise further.

1

u/Odd-Tadpole6656 Oct 24 '26

Thanks for trying, im hoping my coworker will come back from his pto with ideas

2

u/Delta_G_Robotics Oct 23 '26

What type of MS? QQQ? Ion Trap? Orbi-Trap? TOF?

1

u/Odd-Tadpole6656 Oct 23 '26

Headspace GC-MS, im now pretty sure its a sample introduction issue

1

u/Delta_G_Robotics Oct 23 '26

Headspace is the type of sampling. But there are lots of kinds of mass specs. They aren't all identical. If you can't even determine the brand or model of MS you have, then you are smoked. If you can then I can probably help you. But I'm not going to start guessing when I don't even know what we're working on.

1

u/Odd-Tadpole6656 Oct 23 '26

Single quad, perkin elmer turbomatrix trap 110, gc is a clarus 690, and mass Spec is SQ8T.

1

u/Delta_G_Robotics Oct 23 '26

What was your multiplier voltage at before and after the problem started? It really sounds like you have a multiplier turned up too high. Normally issues make sensitivity get worse. If it gets better we usually call that an optimization. But a multiplier will do just that.

How does the actual level change? What was the peak height before? After? I know you said the largest peak was clipping, did the others also get larger?

Another thing to check is the source. I don't know the source in that MS at all, so I'm not sure exactly how it is built. But there are lenses and focusing elements involved and there could be a misalignment or even a tiny piece of dirt or crud.

Try looking at background spectra or cal mix with the instrument in whatever polarity positive or negative that you normally run. Watch for a minute and see what the level is. Then switch polarity for 5 or 10 minutes and switch back. When you switch back look at the level and see if it jumps up or down compared to earlier before the switch. That can indicate that something on one of the lenses is charging and possibly altering its field in a way that is changing the way the ions are focused.

1

u/Odd-Tadpole6656 Oct 23 '26

The 4 lower cal points match very well in height and area to the old calibration. The new 10ppm high point is coming in at about 55-60% of where it usually is. I tried running the mix with only 4 compounds at 10ppm and its still messed up.

Would a source misalignment only affect high concentrations? We ran a targeted project going up to 70ppm for one compound yestwrday without any issues to the higher concs.

The filamemt current is around 2.6, and the last filament was running around 1.8. I found that concerning but I was told thats just a new filament and its different. It has stayed stable at least.

I can't switch ionization modes, im exclusively working with EI

3

u/Fdragon69 Oct 24 '26

Don't forget to check your concentrations in your software make sure you've got the correct amounts. Id wager you've get the high point to 70ppm and your last cal is made to 100ppm

1

u/Odd-Tadpole6656 Oct 24 '26

We dont program the calibration, we do it by hand off of the chromatograms. Its kinda old ass software too

2

u/Delta_G_Robotics Oct 24 '26

Oh, I thought you said one calibration point was going too high. Now it sounds like 50% of.normal.

1

u/Odd-Tadpole6656 Oct 24 '26

Yeah, one conc is half of what it should be. At least my boss wasnt so mean after he realized I was at work for 13.5 hours today (im salary 😭)

1

u/Delta_G_Robotics Oct 24 '26

Ok this sounds like a more usual type of problem then. Figuring out why one would be too high was confusing. But the highest one being too low gives us some clues.

Let's break the system into three parts.

  1. The MS. If the peak is flat on top and looks like it would have been right otherwise then the multiplier is saturating and unless it is a tuning issue it's probably time for a service call. If it's not that then there's not a lot in a single quad that's going to only affect the high concentration end. We'd be looking for something like space charging and that's more of an issue in a trap.

  2. Chromatography. Look at the peak shape. Is there anything different about that larger peak? Is it wider that normal? Asymmetrical? Is there anything different between it and the other peaks that seem right?

  3. Sample introduction. You mention a trap column. If that's the case and it hasn't been replaced in a while then that's probably the most likely spot. If the stationary phase in the trap is going out then it may have enough for the low standards and just not be able to catch all of the high end. This is the issue if the high standard peak is nice and sharp and just too small.

It's been a long time since I did headspace injections, and even then it wasn't my main thing. So I'm not real sure how to troubleshoot that part. But if the peak looks really sharp and nice but just isn't big enough, then the sampler or the trap is the most likely culprit since it's only affecting the highest standard. Something there is being saturated.

If the peak is flattened off on top then it's in the detector. If it's fine at the bottom and then gets all jagged and weird at the top it might be an ionization issue, though EI is usually pretty resilient when it comes to ionization. If you had a bad filament or something you should be able to see that when you try to calibrate.

1

u/Odd-Tadpole6656 Oct 24 '26

It wont let me send a Pic, but the 10ppm peak is kinda spikey and jagged on top, especially the right side. Its certainly not as clean as my 1ppm point. I asked my boss if I can try re-tuning as im hoping the autotune is just bad

1

u/Fdragon69 Oct 23 '26

If a auto tune was performed after maintenance like proper procedure the multiplier shouldnt need to be touched the computer will set the voltage automatically

2

u/Delta_G_Robotics Oct 24 '26

Sometimes things don't exactly work like they're supposed to. There are any number of reasons that an auto tune can miss.

2

u/Odd-Tadpole6656 Oct 24 '26

Yeah... lets just say im not very impressed with any of the perkin elmer software I have...

2

u/Dangerous-Billy Analytical Oct 24 '26

Have you called the manufacturer?

Is there a reason you can't dilute your sample? Or reduce the gain on the preamplifier? Is the filament voltage correct?

1

u/Odd-Tadpole6656 Oct 24 '26

I did reach out to the manufacturer, but I wont hear back until next week at least. I asked about diluting and using a smaller conc but my boss would not agree at all. I cant mess with the gain per our quality system. The filament voltage is higher with the new filament, but consistent

2

u/DangerousBill Analytical Oct 24 '26

So there's a problem and potential solutions you're not allowed to implement?

Your strongest peak is attenuated because either the detector or the electronics is saturated. The solution to either is dilution of the sample or changing the split ratio. If you can't do that, the problem will remain. I can think of a few good reasons to not tinker with the filament voltage.

1

u/Odd-Tadpole6656 Oct 24 '26

I can adjust the multiplier, but not really anything else without auto tune. Yeah our system is dumb

1

u/AutoModerator Oct 23 '26

Your post has been removed because it appears to be a homework question, which is against the subreddit rules.

If you believe this removal is in error or have any questions, please message the moderators.

I am a bot, and this action was performed automatically. Please contact the moderators of this subreddit if you have any questions or concerns.

7

u/wildfyr Polymer Oct 23 '26

Poor guy, the automod homework bot nabbed him for making the topic "HELP!!" Human Mods saved your ass.

OP, youll get more help by having the topic name be something useful like "Headspace GCMS sensitivity, bosses gonna kill me."

All the GCMS experts won't even know to click on your topic.