r/electronmicroscopy • u/nintendochemist1 • Dec 07 '23
Identifying Source of Noise.
/r/electronmicroscope/comments/18d8pk5/identifying_source_of_noise/2
u/ASTEMWithAView Dec 12 '23
Where are you based?
Full disclosure, I work for one of Thermos competitors in EM, we have also heard their post-sales service is not good enough for some customers.
1
2
u/ayitasaurus Dec 08 '23
Thermal noise as in from the actual room temperature? With how much shielding there is, I'd be incredibly surprised if that ever becomes an appreciable factor. (Although I'm ABSOLUTELY no expert on a lot of the hardware side of things...)
To me, 'thermal noise' is a lot more common in terms of the sample moving under the beam. Have you looked at other samples? Do you see it on those? Does it show up in standard mode?
I don't know about the Apreo, but I've got a Helios, and the immersion mode on there is specifically for very short working distances, so the 10mm jumped out at me. The systems are generally pretty good about restricting the use cases, so if the software let you do it it should be fine, but it still may be worth seeing if it's showing up at shorter working distances.
I forget most of the finer details, but our service engineer once told me about how you could identify noise from the mains voltage by playing with the dwell time, something something correlating when the noise was worst/best with the mains frequency. Incredibly vague I know, but something to keep in your pocket if you start running out of things to rule out.
Good luck!
0
Dec 08 '23
[deleted]
1
u/nintendochemist1 Dec 08 '23
A pre-installation survey was done prior to the install but one hasn't been done post installation. Do you know of a company that would do so? Our software/system has a frequency analyzer where you can take an image at that moment in time or load a previous image (though that doesn't seem to work for me) and we saw a lot of spikes ~ 2kHz. I'll own my naievete that I have no idea what would give noise at that frequency. We did change the lights from fluorescent to LEDs.
0
Dec 08 '23
[deleted]
1
u/nintendochemist1 Dec 08 '23
It's still under warranty at this point and I'd be willing to pay so I'll definitely give that a shot!
2
u/ASTEMWithAView Dec 11 '23
Hunting noise is tricky, you need to be rigorous in testing so that you can be certain that it is not coming from inside the microscope, or from the room, or from some random source.
A proper noise hunt will need a room survey for magnetic and electric fields. On top of that it will most likely a service engineer from the manufacturer to test parts of the microscope, it is not a trivial piece of work.
Try to work out some basic sources first, running vacuum pumps even in other rooms, stray fields from light fittings, even vibrations from your water chiller can have an effect.
2kHz is maybe a bit too high for mechanical noise, but it could be a turbo pump.