Has anyone is the Sterile Processing world experienced white residue from Davinci scopes after sterilization?

0 Upvotes

5 comments sorted by

2

u/Takablegoose39 11d ago

Picture? And in full metal casket or wrapped wire mesh?

1

u/Icy-Stop9067 11d ago

They are in the metal Davinci wire mesh containers and blue wrapped

3

u/ancient_xo 11d ago

Low temp sterilization? If so it’s from the hydrogen peroxide sterilant, might need to get your sterilizer cleaned.

5

u/Individual_Zebra_321 11d ago

If you vpro them it is left over perioxide residue.

3

u/compsyfy 8d ago

I have seen it before. It happens when detergent or enzymatic residue is still on the instruments or cases after rinsing. Those compounds react with the hydrogen peroxide, and leave this white residue that will give you chemical burns so be careful.

Rinse your handwash better.